Temp sensor Emotion 6 Detector.

Tags: CT, SIEMENS

(66)
(5)

Mr.CT.service

Temp sensor Emotion 6 Detector.

July 07, 2015 08:28

HI everyone,

In Emotion 6 as you know its DAS detector has three temperature sensors. One of them is showing wrong temp values from 0 to 97.8 °C ramdomly during a scan.
I do not like to replace whole detector just for this sensor. I have thick rings on images.

Sensor1=24.5/Sensor2=0.0/Sensor3=24.4
Sensor1=24.5/Sensor2=54.4/Sensor3=24.4
Sensor1=24.5/Sensor2=31.7/Sensor3=24.4
just some values.

Someone have any code number for replace/substitute it? or any other hint is welcome.

Best regards.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: Temp sensor Emotion 6 Detector.

July 08, 2015 07:55

If it happens sometime ramdomly, Please check the sensor cable(not contacting well).
Please do the data restore, see waht happen?
Lindong123@hotmail.com

Log inor Register

to rate and post a comment

(66)
(5)

Mr.CT.service

re: re: Temp sensor Emotion 6 Detector.

July 08, 2015 10:48

I guess is a false contact too.
I have leave sensor disconnected reading always 8 °C.
When I reconnect it back reading are wrong again.
If IRS table were wrong while sensor is disconnected reading were random values too, weren't it?
I like to have a spare in hand to do repair in only one trip.


Best regards.

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: re: Temp sensor Emotion 6 Detector.

July 09, 2015 02:20

You can change the sensor, ad see what happen.
Where is the unit? I can help you with sensor and high level password (level 7) to diagnotic.
lindong123@hotmail.com
Regards

Log inor Register