Gennadios Tsaptsidis

Statim 5000 Problem with Vibrator Pump

October 12, 2013 09:11

Hello,I have a problem with Statim 5000.It's vibrator pump produces intense sound and vibration and it doesn't pull the water in.I cleaned up the pump and everything was ok .I measure with Ohms and it displays 68 KΩ.Is this result normal?I measure the volts on the vibrator while its active and i get 275 Volt.Where do you think is the problem ,on then vibrator pump or the CPU ?The vibrator pump which is on the machine works with 220 volts.All suggestions are appreciated.Kind Regards

Log inor Register

to rate and post a comment

(8)
(2)

alexium

re: Statim 5000 Problem with Vibrator Pump

October 13, 2013 06:31

Hello Gennadios! Read this article:
http://www.home-barista.com/espresso-machines/repairing-ulka-vibratory-pump-t1079.html
http://coffeetime.wikidot.com/vibration-pump-is-it-worn-out

Log inor Register

to rate and post a comment

(34)
(5)

Matt Du Vall

re: Statim 5000 Problem with Vibrator Pump

October 14, 2013 07:28

I had a simular problem. I replaced the air filter and the problem went away.

Log inor Register

to rate and post a comment

(11)

OWALDO QUISPE PEREZ

re: re: Statim 5000 Problem with Vibrator Pump

October 14, 2013 04:57

Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo

Log inor Register

to rate and post a comment

(22)
(31)

bojan berlot

re: Statim 5000 Problem with Vibrator Pump

October 14, 2013 07:35

Dont loose time with the pump. You have proble in non return valve, praticly the small ball at the end of pump.Replace it. But buy ceme, not ulke. It is the same from coffe machines, price aboute 20€

1 person liked this
Log inor Register

to rate and post a comment

(11)

OWALDO QUISPE PEREZ

re: re: Statim 5000 Problem with Vibrator Pump

October 14, 2013 04:57

Good afternoon:
also I have a problem with a model 2000S STATIM I send the error is nr 60 and not indicating any suggestions please or documentation.
regards
Oswaldo

Log inor Register

to rate and post a comment

(22)
(31)

bojan berlot

re: re: re: Statim 5000 Problem with Vibrator Pump

October 15, 2013 05:56

Cycle Fault #60 For a 134°C cycle, the chamber temperature dropped below the sterilization temperature, allowing for measurement error (i.e. Tchm

1 person liked this
Log inor Register

to rate and post a comment