mdastech
Sonosite titan assert code
September 06, 2017 10:24
Hello all,
I am getting the following assert code on a sonosite titan.
PS: (Ok ==m_prefsFile.Seek (offset))
C: 0xc85 (3205)
D: 0x0(0)
Please does anyone know what this means and how to resolve this blue screen of death.
I have done a hard reset to no avail.
I will appreciate any help with this.
to rate and post a comment
Roberto Hernandez
re: Sonosite titan assert code
September 07, 2017 10:35
Hi, my friend. I think the Motherboard has a circuit with some transducer you put before. If the transducer is connected the machine is supposed to show that error, but if you disconnect the transducer the machine doesn't show this error.
Try with another transducer please and if the machine works, never put it back the transducer because has a bad circuit in the PCB.
I hope this helps you.
Best Regards,
to rate and post a comment