Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

siremobil compact errors

(21)
(1)

Claudio Balangione

siremobil compact errors

February 17, 2012 10:02

Need error list about siremobil compact. Are they in the service software?. Thank you. Claudio Balangione

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: siremobil compact errors

February 17, 2012 09:46

Claudio - The errors are explained while running the Service Software.

1 person liked this
Log inor Register

to rate and post a comment

(21)
(1)

Claudio Balangione

Re: siremobil compact errors

February 20, 2012 10:16

Thanks for the kind reply . The problem is I don't have the service software yet.I'm looking for it. Claudio

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

Re: siremobil compact errors

February 21, 2012 12:04

I emailed you the error codes.

Log inor Register

to rate and post a comment

(21)
(1)

Claudio Balangione

Re: siremobil compact errors

February 22, 2012 09:22

Thank you Waine, I've received your email with the error list. The problem with the Memoskop is the power supply. Claudio

Log inor Register

to rate and post a comment

Mohammed Abdullah

re: Re: siremobil compact errors

March 24, 2015 09:35

Dear Sir,

kindly send me a copy from the error list for this machine,

my email

magmesme@gmail.com

regards

Log inor Register

to rate and post a comment

(11)
(1)

Vicent Martinez

re: Re: siremobil compact errors

September 18, 2012 09:01

Please, I need too the error codes. I have a problem with 4021 and 4024 errors. Can you send it to me?
Thank you

Log inor Register

to rate and post a comment

(265)
(50)
(28)

Wayne Horsman

re: re: Re: siremobil compact errors

September 18, 2012 11:21

Vincent -

It looks like a bad Memoskop. email me for a quote.

- Wayne
--------------------------
Wayne M. Horsman, VP
Columbia Imaging, Inc.
wayne@columbiaimaging.com
--------------------------
Error E4021 Memoskop does not answer

Possible trouble: Host repeats a task 3 times, error E4021 is then entered.

Possible help: Check connection cables to the Memoskop.
Possibly Memoskop or D1 defective..



Error E4024 Memoskop initialization timeout

Possible trouble: Host registered a timeout of the Memoskop during the
initialization of the system
Waiting time for the Memoskop reporting exceeded.

Possible help: Check Memoskop.
Check connection cable to the Memoskop.
Possibly Memoskop or D1 defective.

Log inor Register

to rate and post a comment

(121)
(8)
(15)

Imad Muati

Re: siremobil compact errors

February 17, 2012 11:30

please contact imc@muati.com

Log inor Register

to rate and post a comment

(11)
(1)

Vicent Martinez

re: Re: siremobil compact errors

September 18, 2012 09:01

Please, I need the error list, I have a problem with 4021 and 4024 errors.
Can you send me the error list.
Thank you

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI Solutions to Meet Your Team’s Unique Needs

From preventive maintenance through comprehensive cryogen services, BMI-PMI is your trusted partner for new/reconditioned sales, installation, and service of both facility-based and mobile MRI solutions. www.brownsmedicalimaging.com / 800-701-9729

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)