joe ssali

SIEMENS MULTIX SWING POLYDROS IT

December 12, 2016 04:40

I HAVE A SIEMENS MULTIX SWING POLYDROS IT X-RAY MACHINE WITH ERROR 631.

I HAVE CHECKED THE TUBE ROTOR, IT SEEMS TO BE RUNNING WELL. ANY ASSISTANCE?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: SIEMENS MULTIX SWING POLYDROS IT

December 13, 2016 10:10

This is a Tomo issue not a tube rotor issue check for a blown fuse on the D54 board. If you need any more info please send me an email address


Tomo drive. Control error ==> tube rotation motor movement to hard or defective or D54 defective

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: re: SIEMENS MULTIX SWING POLYDROS IT

December 14, 2016 10:31

Thank you Lenny. Very grateful. Do you have the service manual? And do you have access to parts for this machine.

Log inor Register

to rate and post a comment

(121)
(33)
(9)

joe ssali

re: SIEMENS MULTIX SWING POLYDROS IT

December 15, 2016 10:19

Thank you Lenny and Sanchez. The Machine is working now. All because of your support

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: SIEMENS MULTIX SWING POLYDROS IT

December 15, 2016 02:59

I am glad you were able to repair. Take care

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Próximos Eventos de Subasta

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)