Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

Siemens Mobilett XP - Message Error Err37

(10)

Luis Junior

Siemens Mobilett XP - Message Error Err37

February 10, 2021 10:02

The display is showing me Err 37, I would like to know why this error is due. And if anyone has the service manual for this equipment that you can send me.

Log inor Register

to rate and post a comment

(4)

ENRIQUE RIOS

re: Siemens Mobilett XP - Message Error Err37

February 11, 2021 11:31


Greetings Luis, here is the info refering to error 37, good luck.

ERR 37 Error in the kV control loop
System mode:

Exposure Explanation: D916 (CPU) detects an error in the kV control loop.
Action:
· The value KV actual (TP-KV) does not agree with the value KVnominal (TP-KVS).
· Check the cable connections between D962, C1 and the single tank, and between D916 (X8) and the single, D907.
· Check the signal cable between the single tank and device, possibly defective?
· Check the capacitors of the capacitor bank, all LEDs must light on D972 when the MOBILETT XP is "on".

Possible defects: Fuse F1..F12, F14 (80A) on D972; C1..C12 D972, D962; D907; D916; C1 resonance capacitor.