I have finally fixed this problem on my Sequoia C512 and wish to tell you all how I did it in hopes of saving you a big headache when attempting to repair it.

The Sequoia C512 was shutting down hard (would shut off the second you would press the power button) and the screen had a reddish color to it.

I switched the monitor, reset network data, put in new cables, and even switched out the PIC. None of these fixed the red screen issue I was having with this machine.

What DID fix it though was swapping the RDP.
Here's how to access the RDP:
1. Pull the rear RIGHT air filter out.
2. Remove the Philipps screw.
3. Pull off the whole RIGHT panel ( just pop it off)
4. You'll see a metal panel with five screws shutting it. (It'll also have a yellow caution label on the right side of it). Unscrew all the screws and take the panel off to reveal all of the boards.
NOTE: Make sure the machine is turned off and you have grounded yourself (just touch the metal of the machine)
5. The RDP board is on the right-hand side (they're labeled under the boards) and is taken out by lifting the little things on the top and bottom of the board.
6. Either replace it with a new one or take it out of another machine.
Note: If that doesn't work, then replace the system monitor cable. If that doesn't work, replace the monitor. If that doesn't work, replace the PIC.

I'm by no means a service tech in any way. This is how I ended up fixing mine after about 50 hours of troubleshooting. Any color issues such as not seeing any red colors or any other color are usually from a damaged system monitor cable or the monitor itself. The best way to check is to connect the S-VHS cable for the printer to an old monitor to check.

If you have any questions, shoot me an email, and I can try my best to help you out!

Hope this helps.

sam@sonosui.com

1 person liked this
Log inor Register

to rate and post a comment

(192)
(45)
(1)

Dante Nuevo

re: Sequioa C512 Red Screen Fix

July 13, 2015 10:23

Thank you for sharing.

Log inor Register

to rate and post a comment

(9)

soundultra

re: Sequioa C512 Red Screen Fix

December 26, 2018 08:30

I have the same issue and just replaced the RDP board but still not working. Were is the monitor cable located?

Log inor Register

to rate and post a comment

(13)
(1)

karen bello

re: Sequioa C512 Red Screen Fix

August 04, 2025 11:17

Can u assist as my
Unit is not working

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)