Select SP problem

(63)
(11)

Ronnie Aguiar

Select SP problem

October 01, 2009 04:30

heat exchange is off when must on,This error appears when shooting with the following parameters 120 kv 60 ma time 2 seconds, is there any test that can be performed to detect this problem, many thanks for help
erros:

hv tolerance, feedback differs from reference

ec tolerance, feedback differs from reference


heat exchange is off when must be on

errors appear during shooting by SCU, moves in the invert switc in anode left or right the mistake of HEAT EXCHANGE disappears and appears a new error SCAN UBNORMALLY STOPED (ER80, 15,543 FROM SCU) there is some tests to identify problem

help -me please

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Select SP problem

October 02, 2009 09:30

Hi,
This is HV system problem (mostly)
(0) you can use a scu_diag application to determine what exactly is a problem (remember to change the 'diag' switch on RMS to allow scanning
Possible:
(1) HV spike. In this case inspect the HV connectors on the inverters and XRT. Clean up the plugs and check if there are do damage in the sockets. Reconnect with silicone oil
(2) One of the inverters. Use a 'scu_diag' to determine which one. (make a default scan and using view options find the data - voltages and currents). Old invetrers can have a problem with burnt path on PCB (under big capacitors. Then you can feel the smell- especially after long scans.
(3) AMS - Anode Motor Starter. Just replace.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Cyclic Materials - Sell us your permanent magnet MRIs! WE BUY WARM MAGNETS FOR CASH

Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (501 lots)
Cbrx2sales
TIMED
Jul 15, 2026 (124 lots)
PBC Medical Solutions
TIMED
Jul 17, 2026 (11 lots)
C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (136 lots)