Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Scan Hardware stop scan

Tags: CT

(12)
(1)

steven

Scan Hardware stop scan

October 20, 2017 10:02

My Ct is GE Lightspeed 8 slice (SGI computer)
Now every tim try scan.....shower Scan Hardware stop scan
Run "High Voltage Functional " is OK
But Run "Diagnostic Data Collection"
In "Static X-ray off" is pass
In "static x-ray on" will stop scan
error message is 184154
Host:STC
Function:XRAY GENERATION:FinalnExposure Command
Time out on Exposure Command to X-ray on.
HW Error:Possibile hardware problem
Timeout is:300 mS

Has any one can give me suggest.
thank
steven

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Scan Hardware stop scan

October 21, 2017 10:18

There must be a set of error, look at the first. The error you described is a result of other error most probably.
tibor@ivanszky.com

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: Scan Hardware stop scan

October 23, 2017 09:09

It can be problem with tube
How mane mas is done ?

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

re: Scan Hardware stop scan

October 23, 2017 09:09

Hello
Could you make Again x Ray in DDC, then click on last page and take picture to last 5 pages and send me
Raúl
raul.elizalde.sti@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(12)
(1)

steven

re: re: Scan Hardware stop scan

October 24, 2017 09:41

run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass

thank for your message
steven

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Scan Hardware stop scan

October 23, 2017 09:09

contact me craig@telemed-technologies.com and we work on this together

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Scan Hardware stop scan

October 23, 2017 06:06

Check fiber optics lines from console to DAS. Be careful, visible light pass test does not work.
You must use infrared test. TV remote control and old mobile phones with IR sense cameras work only. Two person Job.
The most probable optical cable fault is on the rotating part. Try stationary scan, gantry rotation power off and connect optical cable directly from Console to DAS.

1 person liked this
Log inor Register

to rate and post a comment

(12)
(1)

steven

re: Scan Hardware stop scan

October 24, 2017 09:41

run DDC static x-ray off is pass
static x-ray on ......stop scan
But We found in Flash download
Update .....can finish and ne error show
Querying....will show STC unknow,ETC unknow,OBC unknow
My friend talk to me ,try load old states,but no help
We try Rotor Control.......pass

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: Scan Hardware stop scan

October 24, 2017 10:21

Check all fuses and boards in HEMRC assembly...and perform rotor diagnostic tests....
Check the OBC-HEMRC Interface board
Check the KV & MA Boards in the OBC...If possible swap the Artesyn II boards between the STC and OBC (mind the jumper settings on each)

Log inor Register

to rate and post a comment

(9)
(2)

Georgi Gaydov

re: Scan Hardware stop scan

September 16, 2018 02:47

Hi Steven,
I have the same problem.
How you solve your problem.
Thanks
Georgi

Log inor Register

to rate and post a comment

(9)
(2)

Georgi Gaydov

re: Scan Hardware stop scan

September 16, 2018 02:50

Hi Steven,
I have the same problem.
How you solve your problem.
Thanks
Georgi

Log inor Register

to rate and post a comment

stats
DOTmed text ad

PHILIPS MX8000 & Brilliance Systems & Spare Parts In Stock Now!

HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com

DOTmed Certified stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)