Prospeed ERROR 117/153

July 25, 2008 09:55

I need help on a GE Pro speed system and Im wondering if you can help. I'm getting an error 117/153(filament related) and I have tried the CPU board but still no go..any assistance(ASAP) would be great.

given:

Tube filament ohms OK
tried Cathode cable and Fil. board..no go
CPU board was replace which cause the problem to occur aafter 2 exposures instead of instantly.??

Log inor Register

to rate and post a comment

(49)

Kakhaber Gvaramia

Re: Prospeed ERROR 117/153

July 25, 2008 03:31

check the LEDs on filament board. did you ran service tests S21-S24?

Log inor Register

to rate and post a comment

(317)
(22)
(8)

Joseph Jenkins

Re: Prospeed ERROR 117/153

July 26, 2008 09:26

I found the tube was bad..Thanks

Log inor Register

to rate and post a comment