picker explorer II error E97

Tags: Radiology - R/F, PICKER

(174)
(8)
(31)

Hipolito Jorge Iglesias

picker explorer II error E97

July 08, 2014 10:13

Dear Colleagues,
i am triying to get some help with a problem whith this machine, i have replaced the EEprom IC26 of the microprocessor PCB because i was getting some errors, now when i do an exposure with some factors i got the error E97, it says an overlong exposure but i measured the mAs and KVp values and were ok, with other values i do not have any error 5 mAS and 65KVP for example, i do not know if the machine needst o be recalibrated and in this case i would like to know how to access the calibration procedure.
Thank you
annorys1971@att.net.mx

Log inor Register

to rate and post a comment

(71)
(13)

gerardo millena

re: picker explorer II error E97

July 09, 2014 09:50

Does it have an optional AEC. This could be active, try to disable this. Good luck

Log inor Register

to rate and post a comment

(174)
(8)
(31)

Hipolito Jorge Iglesias

re: re: picker explorer II error E97

July 10, 2014 04:42

Dear Gerardo
It does not have aec option, i have no idea how to disable it in case that it is programed. the machine have all the apr active. but the error happens even in manual mode.
thanks

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

re: picker explorer II error E97

July 09, 2014 09:51

my friend really you need to calibrate the machine beause some of parameter was change but is easy to do this only you need to makesure that the heating current be set.

mailme at xelectjg@yahoo.com for send you the instructions

Log inor Register

to rate and post a comment

(174)
(8)
(31)

Hipolito Jorge Iglesias

re: re: picker explorer II error E97

July 10, 2014 04:43

Dear Joel,
yes i think it needs to be recalibrated the machine because i have swapped several Printed circuit boards of another machine, for recalibration is it necesary to use the program keyswitch with steps 1,2 and 3 or just to acces the the process by means of the switch 2 to get access the level 2 on the system?
Thank you for your help

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Próximos Eventos de Subasta

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)