Philips-BV Endura M050 Error Code

Eduardo Tognon

Philips-BV Endura M050 Error Code

February 05, 2011 05:11

I am getting the following error when booting up the tower "Could not
start application, because of application software inconsistency".
With C-Arm connected I get an M050 error on it but i belive it is caused by the DFI software not booting up on the tower. (DFI software version 2.2.4 data 23 july 2007).
The problem is software?
I need the procedure reloading software and service manual.
Any info would be
greatly appreciated.
Eduardo
BRAZIL
eduardo@elomed.com.br

Log inor Register

to rate and post a comment

(57)

LUIS RIVAS

re: Philips-BV Endura M050 Error Code

July 02, 2013 06:01

hello you solve your problem with endura error m050, I have the same error, can you help me

Log inor Register

to rate and post a comment

(57)

LUIS RIVAS

re: re: Philips-BV Endura M050 Error Code

September 04, 2014 10:18

Yes was the coaxil cable within the cable to connect the trolley with stand

Log inor Register

to rate and post a comment

(2)

Miguel Torres

re: re: Philips-BV Endura M050 Error Code

March 22, 2019 10:44

I have the same error in pulsera c arm could you help me
Wath means this error?!

Log inor Register

to rate and post a comment

(23)
(6)
(64)

G. John Oggel

re: Philips-BV Endura M050 Error Code

September 03, 2014 09:00

Hello,
did you figure out how to fix this error code? I have a philips BV-Endura that has the same code...can you help me?

1 person liked this
Log inor Register

to rate and post a comment

(57)

LUIS RIVAS

re: re: Philips-BV Endura M050 Error Code

September 05, 2014 09:38

The coaxial pins in the stand connector were in short, eliminate this short circuit and the fail dissappears then change the cable, is expensive but is the solutions

Log inor Register

to rate and post a comment

(53)
(10)

hiortsam dade

re: Philips-BV Endura M050 Error Code

January 11, 2019 09:51

Please whats the part number of this cable?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Próximos Eventos de Subasta

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)