Over 300 Virginia Auctions End Today - Bid Now
Over 300 Total Lots Up For Auction at Three Locations - Over 100 California Auctions End Tomorrow 07/22 - Bid Now, KS 07/24, CA 07/31

phhilips tomoscan CX

sunil Kumar

phhilips tomoscan CX

August 19, 2004 11:16

i suddenly got error X26 during scanning which went of after cleaning of HT cable plugs and tube which were dry and had some deposit on it.now , after a week it(X26) has started coming again .Is it defective tube or can be something else

Log inor Register

to rate and post a comment

(14)

(19)

Bill Blackford

Re: phhilips tomoscan CX

August 19, 2004 01:06

Did you put new grease on the HT plugs.
we use DOW Croning #4.
If you don't do this you will get the same problem.
If you need a tube I have a Gs 1580 Eimac tube
in a B180 Housing.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

We repair MRI Coils, RF amplifiers, Gradient Amplifiers and Injectors.

MIT labs, experts in Multi-Vendor component level repair of: MRI Coils, RF amplifiers, Gradient Amplifiers Contrast Media Injectors. System repairs, sub-assembly repairs, component level repairs, refurbish/calibrate. info@mitlabsusa.com/+1 (305) 470-8013

stats

Próximos Eventos de Subasta

C2 Management
TIMED Ending Soon
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (229 lots)