MX8000 start up problem

April 24, 2013 10:19

During system start up on MX8000 CT, Gantry can not turn on, it turns on but immediately turns off several times at the end keeps off
on Error Log
Even Log: E CT_MCU_28 Control Error MCU

I already checked control signal on slipring but looks ok
any idea?
Thank you

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: MX8000 start up problem

April 24, 2013 06:00

Check the e-stops. Check the safety switchs.
Check at the MCU there is no in service mode, and the manual brakes are off.
Good luck!

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

re: re: MX8000 start up problem

April 25, 2013 12:02

Operator reports a sound on PDC, during Scan , and I just found F8 tripped. I reset it, but Gantry does not turn on yet, I already checked e-sw, MCU, are ok

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: re: re: MX8000 start up problem

April 25, 2013 07:55

Check the switchs a side the MCU. There are a few switchs thats manage the heat exchangers fan, the dehumidifer, etc.

Log inor Register

to rate and post a comment

(45)
(1)

chalie alvarez

re: MX8000 start up problem

April 26, 2013 09:51

what is the error thst you hae at the error log, this is important, this means that the mcu has find some problem on the machine can be some thing simple or a board, fan not working, where are you located???

Log inor Register

to rate and post a comment

(45)
(1)

chalie alvarez

re: MX8000 start up problem

April 26, 2013 09:51

im located in texas, the best and faster way is to conect the computer to the mcu and will read the exact error and fix it fast send me email

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

re: re: MX8000 start up problem

April 26, 2013 01:50

SYNC ERR red led ON, on MCU board
Erroo Log
E CT_MCU_28 Control Error MCU

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: re: re: MX8000 start up problem

April 26, 2013 06:27

Please, copy the complete error code, like E MCU 00 XX XX XX, so I can give you more information.
The sync problem is a product of the incomplete boot.

Log inor Register

to rate and post a comment

(128)
(10)
(5)

Manuel Esquer

re: re: re: re: MX8000 start up problem

April 27, 2013 08:00

Comp MCU
Sevir CRIT
Code 28
Message E 0 0 1c 3 e9 0 0

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: re: re: re: re: MX8000 start up problem

April 28, 2013 10:25

The control signals from/to the rotating part of the gantry are transferred
via a slipring interface.
Noisy or defective data transfer from ACU D513 to MCU D301. Noisy environment. EMC interference.
Please, check and Test signal path and slipring interface with control data transmission test
Chcek control data slip ring brushes
Clean control data slip ring
Replace ACU
Replace MCB2 on ACU
Replace MCB2 on MCU
Replace MCU