Sang-Yoon Park

MX8000 Error

April 18, 2012 10:08

Dear Sir,

When we power on, during the booting, Error display “Master Server failed-MCU Reset Failed” in MX8000 which is Philips CT.

Please help me.

Log inor Register

to rate and post a comment

Yuriy Kuznetsov

Re: MX8000 Error

April 18, 2012 01:56

at first, write your e-mail

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

Re: MX8000 Error

April 18, 2012 06:28

Could be anything. Try to check the Error Log, filter for CRITICAL and write here the error code.
Check the Gantry is POWER ON, check the IRS is ON.
Also, try to check any Interlock on the Gantry. If any interlock is open, the MCU will not reset.
Regards,

Log inor Register

to rate and post a comment

(15)
(1)
(1)

Eduardo Arr

re: Re: MX8000 Error

December 05, 2012 10:07

hello Gustavo:
What do you mean about the interlock, the e stops?

Log inor Register

to rate and post a comment

(107)
(11)
(3)

Gustavo Volpi

re: re: Re: MX8000 Error

December 05, 2012 06:38

Yes, all the Stop-Reports have a e-stop or interlock. Something is avoiding to the Gantry to Start Up.
You can check the stop report at the MCU, with the test points, or from the operator´s console, some times, the error log shows you why the MCU can´t be reseted. Or you can connect the CANPro to the system and check them out.
There is a lot of options to figure out what´s going out.
Good luck!

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: re: Re: MX8000 Error

December 06, 2012 09:45

I have the service manual for the 8000 e-mail me crisb1955@aol.com

Log inor Register

to rate and post a comment

(30)
(6)

Fernando Gomez

Re: MX8000 Error

April 18, 2012 08:28

Do you know how to read the error log?

Log inor Register

to rate and post a comment

junn refol

re: Re: MX8000 Error

September 11, 2014 12:55

Check the slip ring and the data brushes then check the rotation module...

Log inor Register

to rate and post a comment