Over 500 Total Lots Up For Auction at One Location - NJ 06/15

MVME162-510A NVRAM E Positioner help needed

Tags: Cardiac Imaging

(1)

MIGUEL ANGEL ROCHA

MVME162-510A NVRAM E Positioner help needed

November 01, 2016 10:56

Dear seniors,
As my Positioner NVRAM battery expired, i changed it with new one the i flash it with my CPU firmware Figaro 7.1
my target is MVME 162-150A
My email k-okos@hotmail.es
Thank you

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: MVME162-510A NVRAM E Positioner help needed

November 03, 2016 10:38

Hi,

At first you should Configure your MVME board and load the new NVRAM. (Once the Battery fails the board will be set to default. You will have to set it according to the GE Service manual)
Secondly you should Calibrate the Positioner and load the settings to the NVRAM.

R.Gopal
aparnaelectromech@gmail.com

Log inor Register

to rate and post a comment

(14)
(2)

Sergiy Shurayev

re: re: MVME162-510A NVRAM E Positioner help needed

May 11, 2017 09:17

Hi,
could you tell me, please Direction # of GE manual you talked about.
BR,
Sergiy Shurayev.

Log inor Register