Over 800 Total Lots Up For Auction at Four Locations - VA 07/21, CA 07/22, KS 07/24, CA 07/31

Indico 100 generator

(91)
(38)

Kazem Saremi

Indico 100 generator

December 07, 2012 10:08

Dear friends
Who is expert on CPI Indico 100? When I push the Fluoro foot switch, the rotor will run but there is no fluroscopy and the fluoro timer is still zero. After releasing the foot switch the rotor is still running until I change the post (from fluoro table to bucky stand or Aux)
What is the problem
Best Regards
Kazem
mk.saremi@gmail.com

1 person liked this
Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

re: Indico 100 generator

December 07, 2012 11:30

is it coupled to a digital system or conventional RF system
try to check the II amplifier power supply or a video chain
also , may be you lost the fluoro configuration in gen.

1 person liked this
Log inor Register

to rate and post a comment

(91)
(38)

Kazem Saremi

re: re: Indico 100 generator

December 08, 2012 11:18

Dear Tahar
This generator is coupled to a conventional RF table (made by GMM for BMI - BRT 20). I found the fault in video chain and removed it. Now I have fluoroscopy . But after releasing the foot switch, there is no X-ray but the rotor is still running. In order to stop the rotor running,I change the workstation (from fluoro to Rad).Let me know what is the problem.

1 person liked this
Log inor Register

to rate and post a comment

(47)
(7)
(6)

KONSTANTINOS PATTICHIS

re: re: re: Indico 100 generator

December 08, 2012 07:43

You have configuration problem on the generator, follow the service manual.

Log inor Register

to rate and post a comment

(80)
(23)

Fysares Delali

re: Indico 100 generator

December 09, 2012 01:55

Dear Kazem, please send me e-mail. I have configration menu settings...fysares@hotmail.com

Log inor Register

to rate and post a comment

(9)

rsimaging

re: Indico 100 generator

December 09, 2012 05:00

Please check your configuration...I would look at your fluro hang time in setup..

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Red Kingdom Wholesale Surplus Auctions July 24th

Red Kingdom Wholesale closeout auctions take place on DOTmed.com. Bid today on 115 lots of medical equipment. Onsite or virtual inspection available.

stats

Próximos Eventos de Subasta

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)