Hello:
I am having a TGGC Fault 2100 X Ray Tube Current without Firmware request during Tube Warm up and scans. Any ideas about possible causes? Thank you.
Best Regards,
Elias Valdes
ervaldes@att.net.mx
Check the adjustment procedures for the A2 and A3 boards in the TGGC. A maladjusted board can cause this as well as a faulty A2 or A3 board. Make sure the X-Ray On adjustment on page 6-4-1 of the X-Ray manual is ok as well. Also, check and/or replace the precontacting and contacting SCR's on the XG1A6 Panel. A HV Transformer can also cause this problem.
The info from Eliott is correct. However, with over 25 years of experience on this TGGC system, odds are your tube is about to go out. Clean the cables and tube wells. If you hurry, you may be able to have the tube reprocessed and save it for a while.
I agree with Scooter.... more than likely it's the tube, but be sure to check out the other things I posted first... a small adjustment is cheaper than an X-Ray tube
Hello:
Thank you both of you. I will recheck the adjustments as I had already gone through them, but I will do it again to make sure everything is OK. Actually the system originally was a 9800 and it was upgraded to HiLight Adavantage. I have another question. In TGGC CPU board there seem to be missing a couple of EEPROM's, I wonder if they are used or not, mainly one of them, U176. I don't know for sure it is needed. Thank you in advance.
Best Regards,
Elias Valdes
U176 is the select 15 eeprom and certainly should be there. Be careful that the revision level of the replacement prom is the same. Most later systems are running either rev 0 or rev 12. They all must match revision levels.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Elias Valdes
HiLight Adv Error
April 10, 2007 11:05
Hello:
I am having a TGGC Fault 2100 X Ray Tube Current without Firmware request during Tube Warm up and scans. Any ideas about possible causes? Thank you.
Best Regards,
Elias Valdes
ervaldes@att.net.mx
to rate and post a comment
Elliot Marcus
Re: HiLight Adv Error
April 10, 2007 04:58
Check the adjustment procedures for the A2 and A3 boards in the TGGC. A maladjusted board can cause this as well as a faulty A2 or A3 board. Make sure the X-Ray On adjustment on page 6-4-1 of the X-Ray manual is ok as well. Also, check and/or replace the precontacting and contacting SCR's on the XG1A6 Panel. A HV Transformer can also cause this problem.
to rate and post a comment
Ralph Childs
Re: HiLight Adv Error
April 11, 2007 09:37
Elias,
The info from Eliott is correct. However, with over 25 years of experience on this TGGC system, odds are your tube is about to go out. Clean the cables and tube wells. If you hurry, you may be able to have the tube reprocessed and save it for a while.
Scooter
scanner1@eim.ae
to rate and post a comment
Elliot Marcus
Re: HiLight Adv Error
April 11, 2007 03:03
I agree with Scooter.... more than likely it's the tube, but be sure to check out the other things I posted first... a small adjustment is cheaper than an X-Ray tube
to rate and post a comment
Elias Valdes
Re: HiLight Adv Error
April 13, 2007 03:04
Hello:
Thank you both of you. I will recheck the adjustments as I had already gone through them, but I will do it again to make sure everything is OK. Actually the system originally was a 9800 and it was upgraded to HiLight Adavantage. I have another question. In TGGC CPU board there seem to be missing a couple of EEPROM's, I wonder if they are used or not, mainly one of them, U176. I don't know for sure it is needed. Thank you in advance.
Best Regards,
Elias Valdes
to rate and post a comment
Ralph Childs
Re: HiLight Adv Error
April 14, 2007 12:18
Elias,
U176 is the select 15 eeprom and certainly should be there. Be careful that the revision level of the replacement prom is the same. Most later systems are running either rev 0 or rev 12. They all must match revision levels.
Scooter
scanner1@eim.ae
to rate and post a comment