One is a CTI which has a MX 200 tube that just went bad. My question is what needs to be done to convert this unit to a MX 165 tube; other than balance it and changing the settings on reconfig?
The Second unit I have is a Synergy; this unit is giving me a
problem with the database, I can warm up the unit, calibrate it and
it works fine, but when I make a scans it does not display the
images, (scout or axial). when I do and axial it gives me the following error:
application unexpectedly terminated: recon
the system has been shutdown
try to restarting the system by pressing "start up"
I try to regenerate the database by initializing the "sdcidbm"
directory but with this unit is not working. I don't have the cd
driver to reload software; and I think this is probably my next
step. Any suggestions?
I'm pretty sure that's all you need to do to downgrade the tube type to the MX165 (rebalance the gantry and change the tube type in reconfig). If I am wrong, I am sure someone here will correct me.
As far as the Synergy goes... my guess us that you either have corrupted software or a bad hard disk. Either way you will need to reload software in my opinion.
Ernesto Guerra
help with ge cts
July 09, 2008 11:18
I'm working in Mexico on two units.
One is a CTI which has a MX 200 tube that just went bad. My question is what needs to be done to convert this unit to a MX 165 tube; other than balance it and changing the settings on reconfig?
The Second unit I have is a Synergy; this unit is giving me a
problem with the database, I can warm up the unit, calibrate it and
it works fine, but when I make a scans it does not display the
images, (scout or axial). when I do and axial it gives me the following error:
application unexpectedly terminated: recon
the system has been shutdown
try to restarting the system by pressing "start up"
I try to regenerate the database by initializing the "sdcidbm"
directory but with this unit is not working. I don't have the cd
driver to reload software; and I think this is probably my next
step. Any suggestions?
to rate and post a comment
Elliot Marcus
Re: help with ge cts
July 10, 2008 08:13
Ernesto,
I'm pretty sure that's all you need to do to downgrade the tube type to the MX165 (rebalance the gantry and change the tube type in reconfig). If I am wrong, I am sure someone here will correct me.
As far as the Synergy goes... my guess us that you either have corrupted software or a bad hard disk. Either way you will need to reload software in my opinion.
Best of luck,
Elliot
to rate and post a comment
farhan zeb
Re: help with ge cts
September 17, 2011 07:12
INDIGO2 4.4 version
1_) dear friend i have loaded the software first i loaded the OS then
Applications and after that i have to configure or load the SBC with OC when i type the command loadSBC so the
following error occurs Not enough free spcae can not load network software
when i try to connect to SBC i type the command cu sbc but following error occured connect failed: cant't access
device
i have tried all the procedure from LFC but still same problem i have tried to bring it to the MC6080> prompt but
still nothing happened still same problem
but when i use the command rsh sbc and reconfig the sbc and after st so during initialization system goes shut down
itself .
NOTE:
when i have loaded the application software so after that when i open the unix shell so an message
appeared executing install.dasm no dasm's found
so please help me i'll wait for your reply
2_) Dear friend i have lost the system state also of cti indigo2 system and and the oc disk drive have been crashed
i have
replace the disk drive and loaded the software from cold so i have lost my system state so now i have loaded the
software but i have no system state but there is need of the system state so what should i do now there is the
method to restore files in the LFC but how should i access the unix shell with crash disk drive so please tell me
what will be the alternative way to skip this step and start the applications.