Hi everyone,
I have a minor problem with a GE proteus generator after taking an exposure and then removing the cassette from the table bucky the system intermittently plays a tune a series of notes? . Has anyone come across this before . It doesn't seem to affect system operation.
That sounds like the bucky drive motor that has a delay on driving the grid back to home position after the exposure. It may eventually fail and the generator won't expose on that bucky.
check the power supplies on the system I/F board in the top of the Jedi cabinet if the the +5 VDC is at or close to 4.85VDC in the base of there is a power supply if you look at the wires coming off the supply you will see a connector with power off reseat this connector(if you have contact cleaner spray or apply to both sides) then retest the supplies.
the intermittent tone can be a reset tone due to these supplies
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Michael Kakoulli
Ge proteus
November 26, 2015 10:01
Hi everyone,
I have a minor problem with a GE proteus generator after taking an exposure and then removing the cassette from the table bucky the system intermittently plays a tune a series of notes? . Has anyone come across this before . It doesn't seem to affect system operation.
Regards
Mike
to rate and post a comment
xrayrepairdude
re: Ge proteus
November 27, 2015 08:12
That sounds like the bucky drive motor that has a delay on driving the grid back to home position after the exposure. It may eventually fail and the generator won't expose on that bucky.
Ken
to rate and post a comment
Michael Kakoulli
re: re: Ge proteus
November 27, 2015 10:57
Hi ken,
will check it out.
Mike
to rate and post a comment
Lenny Place
re: Ge proteus
November 27, 2015 03:46
check the power supplies on the system I/F board in the top of the Jedi cabinet if the the +5 VDC is at or close to 4.85VDC in the base of there is a power supply if you look at the wires coming off the supply you will see a connector with power off reseat this connector(if you have contact cleaner spray or apply to both sides) then retest the supplies.
the intermittent tone can be a reset tone due to these supplies
to rate and post a comment
Michael Kakoulli
re: re: Ge proteus
November 28, 2015 07:18
Thanks will check at next service visit.