110-
0903H
70 degree C thermal
safety error
70° loop detected open
1. X-ray tube too hot;
normal error
2. Wiring problem
3. Sensor problem (Tube)
4. rotation board.
IT COULD BE A BROKEN WIRE OR DEFECT TEMP. SENSOR.
CHECK CONTINUITY (OHM)
CHECK ON THIS SITE THERE IS THE SERVICE MANUAL:
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
ge proteus x.ray
July 10, 2015 03:49
I looking for service manual of x.ray machine ge Proteus RX/a or help for error 110,903
to rate and post a comment
Michael Garnica
re: ge proteus x.ray
July 11, 2015 09:22
error code means Tube exceeded 70 degree C it got too hot. Thermal error code and 110 804 Tank thermal error
to rate and post a comment
Wayne Horsman
re: ge proteus x.ray
July 11, 2015 09:22
what is your email address?
to rate and post a comment
Mohammad irfan
re: re: ge proteus x.ray
July 12, 2015 02:20
Irfanrandhawa@yahoo.com
to rate and post a comment
Wayne Horsman
re: ge proteus x.ray
July 11, 2015 09:22
110 => Thermal error, Class 5, Code 0903H, Tube exceeded 70degC
to rate and post a comment
Engr Ali Hussain
re: ge proteus x.ray
July 11, 2015 09:27
Hi
GE have online free library of all manuals. You can download it online..
Just google GE Healthcare manual.
Thank you.
to rate and post a comment
Kamran Butt
re: ge proteus x.ray
July 11, 2015 09:30
Download it from ge support documentation library.,
to rate and post a comment
Arturo Papa
re: ge proteus x.ray
July 11, 2015 09:30
110-
0903H
70 degree C thermal
safety error
70° loop detected open
1. X-ray tube too hot;
normal error
2. Wiring problem
3. Sensor problem (Tube)
4. rotation board.
IT COULD BE A BROKEN WIRE OR DEFECT TEMP. SENSOR.
CHECK CONTINUITY (OHM)
CHECK ON THIS SITE THERE IS THE SERVICE MANUAL:
http://apps.gehealthcare.com/servlet/ClientServlet?REQ=RNEW&MODALITY=XR
to rate and post a comment
Mohammad irfan
re: ge proteus x.ray
July 12, 2015 02:20
Thanks to all thermal sensor and tube temp is normal and wiring OK some time error appeared