I have a Lunar DPX-Bravo with encore 2006. QA does not go through and give an error "Swing arm not locked in scanning position. Please lock before continuing."

I tried unlocking and locking the swing arm several times with no change. Any advice on troubleshooting will be highly appreciated.

Log inor Register

to rate and post a comment

(5)

Juan Gaytan

re: GE LUNAR DPX-BRAVO

April 23, 2019 10:06

you have the service manual? write me
Regards
juan.p.gaytan@gmail.com

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

re: GE LUNAR DPX-BRAVO

April 23, 2019 10:14

Hello Mohamed, you could follow cables trought inside arm to bulk heal connector inside tablet ir do a bypass on switch ti determine if is bad. Feel free to contact me if need It. lemedina1964@yahoo.com.mx

Log inor Register

to rate and post a comment

(408)
(38)
(5)

Juan de la Cruz Martinez

re: GE LUNAR DPX-BRAVO

April 23, 2019 11:13

Have you check if there is a switch that detects the arm is locked?

Log inor Register

to rate and post a comment

(95)
(7)
(2)

Rick Ebling

re: GE LUNAR DPX-BRAVO

April 24, 2019 10:16

There are TWO switches in the rotating vertical arm. They have to both be actuated mechanically.

Log inor Register

to rate and post a comment

(72)
(17)
(7)

Ben Williams

re: GE LUNAR DPX-BRAVO

April 24, 2019 10:17

HI Mohamed,

You may have a bad limit switch in the arm. There is a limit switch up there that will sense the open and closed position. I would first check to ensure that switch is engaging correctly. If you need parts or other advice let us know.

Ben