GE LS16 Pro 5.x

Tags: CT, GE

(3)
(1)

Keith Ader

GE LS16 Pro 5.x

April 21, 2018 01:16

Getting intermittent error on system during scans.

Host:tgp Proc:tgp_subsys Error:260100152
Exception Class:Abort Severity:Primary
File:sclsrService.c Line#:386
Function:System Monitoring :Xray Exposure Interlocks
Scan Type:Helical Scan:7163/2/1
Exposure Interlock Opened.
The current loop from Gantry to PDU.
The cause of the failure: Console Relay

Replaced the PDU Control Board in the NGPDU a week ago. Ran fine until today and got same error. This is intermittent and I can run the system for days without error. Is there another set of relays that I am missing in the NGPDU or are they calling the contactors "relays"? Thoughts?

Log inor Register

to rate and post a comment

(39)
(8)

Raul Elizalde

re: GE LS16 Pro 5.x

April 23, 2018 09:54

Send me your email, I will can send you some information about interlocks

3 people liked this
Log inor Register

to rate and post a comment

Hong Vinh Vo

re: re: GE LS16 Pro 5.x

April 30, 2018 09:42

can you send me about the interlocks information?
My email is vohongvinh@gmail.com.
Thanks alot.

Log inor Register

to rate and post a comment

(3)
(1)

Keith Ader

re: re: GE LS16 Pro 5.x

May 01, 2018 09:46

keith.ader@uky.edu

Log inor Register

to rate and post a comment

Mario1

re: GE LS16 Pro 5.x

April 23, 2018 10:56

Hello,
Please check error log.for another messages,console relay(in DIP board)just stop exposure.
Probably problem with data transfer via optic fiber.

Regards,
Maris

Log inor Register

to rate and post a comment

(3)
(1)

Keith Ader

re: GE LS16 Pro 5.x

May 01, 2018 09:47

I looked at the signals in the documentation I had and it appears that this interlock is the door switch interlock. Since the door switch is not enabled (jumped in the NGPDU) and the NGPDU Control Board was replaced I replaced the DIP (the signals appear to run through there). I have not had an issue with it so far since that replacement. If it pops again I will update this post. Thanks for the thoughts and input.

1 person liked this
Log inor Register

to rate and post a comment

(1)

Rodrigo Coliboro

re: re: GE LS16 Pro 5.x

September 25, 2020 11:10

Hello, my CT brightspeed 4 has the same error, is your CT with this error okay?

Log inor Register

to rate and post a comment