Over 800 Total Lots Up For Auction at Four Locations - VA 07/21, CA 07/22, KS 07/24, CA 07/31

Ge Logiq 400 scsi error ( boot problem)

Tags: Ultrasound

Posted by an Anonymous User

Ge Logiq 400 scsi error ( boot problem)

March 19, 2014 10:04

I have a ge Logiq 400 cl pro series when we switch on the machine machine show SCSI error.machine have 5.01 version

waiting any suggestion about this problem

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: Ge Logiq 400 scsi error ( boot problem)

March 19, 2014 08:58

CAn you give more details on the SCSI error? For the meantime try to check the SCSI cable connection, the power supply for your hard disk. Possible suspect is hard disk drive

1 person liked this
Log inor Register

to rate and post a comment

DigitalUltrasounds

re: re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 08:21

i checked all the things they are ok but when i change the hard drive with other machine.then machine work ok.i need support for solve this problem

Log inor Register

to rate and post a comment

(90)
(19)

bens prado

re: re: re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 08:40

Your Hard disk drive(HDD) is bad. Replace your HDD and load the software. If you dont have the software copy the HDD of the other machine HDD to HDD copying.Bring it to the computer shop that is capable of duplicating your entire HDD.

Log inor Register

to rate and post a comment

SHYJU MON

re: re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 08:24

Chek Low voltage power supply Distribution board which is having two fuses,lcated left side of LV powersupply remove fuses and make it connect directly, and see stil not solved your hard disk ,scassy card, scassy cable may be faulty

we have all these parts
shyjumon@hotmail.com

1 person liked this
Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 10:38

send me you e mail for asm manuals

Log inor Register

to rate and post a comment

only rocky

re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 03:01

hii is it scsi error 20166 max problum is scsi card which is below of the HDD, and also scsi cable chick once 5v and 12v in LV

Log inor Register

to rate and post a comment

(97)
(21)
(32)

Zach Ruder

re: Ge Logiq 400 scsi error ( boot problem)

March 20, 2014 06:31

This does not always indicate a bad HDD. It can also be the CPU #12 Board which also has a scsi controller built into it. It can also only be a timekeeper issue as I have seen this before.

Log inor Register

to rate and post a comment

DigitalUltrasounds

re: re: Ge Logiq 400 scsi error ( boot problem)

March 22, 2014 10:52

ok i will check

Log inor Register

to rate and post a comment

(58)
(1)

MOTSO

re: Ge Logiq 400 scsi error ( boot problem)

April 02, 2014 07:44

I HAVE THE SOLUTION OF YOUR TROUBLESHOUTING mail me on theodmotso@gmail.com

Log inor Register

to rate and post a comment

(16)

babusgajelli Gajelli

re: Ge Logiq 400 scsi error ( boot problem)

May 09, 2014 01:54

Pls check all kinds of bus cables&some ribben cables if you are all ready done this ignore it

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Red Kingdom Wholesale Surplus Auctions July 24th

Red Kingdom Wholesale closeout auctions take place on DOTmed.com. Bid today on 115 lots of medical equipment. Onsite or virtual inspection available.

stats
advertisement

Próximos Eventos de Subasta

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (218 lots)