GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM

Fadhl Alakwa

GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM

October 01, 2012 10:06

Hi
My system GE LS.X3 has the following error
"the hardware reset failed". When I replaced the STC ARTESYN III board the error disappear and I got another error when I do KV&MA (X-RAY)TEST:

"Hard time: 86098 error type:firmware software error :error when posting to vrtx queue soft or hard reset may be required."
I appreciate any advise and suggestions.

Note: I do download flash without any errors.
I am waiting your reply.

Log inor Register

to rate and post a comment

Mark Parris

re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM

October 02, 2012 07:24

Did you change the jumpers on the Artesyn bd?

Log inor Register

to rate and post a comment

Fadhl Alakwa

re: re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM

October 04, 2012 08:44

Yes the jumpers are like the service manual configuration. You can not imagine that this error is because 2 broken resistors in the Axial board.
Thank you

Log inor Register

to rate and post a comment