Hi
My system GE LS.X3 has the following error
"the hardware reset failed". When I replaced the STC ARTESYN III board the error disappear and I got another error when I do KV&MA (X-RAY)TEST:
"Hard time: 86098 error type:firmware software error :error when posting to vrtx queue soft or hard reset may be required."
I appreciate any advise and suggestions.
Note: I do download flash without any errors.
I am waiting your reply.
re: re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 04, 2012 08:44
Yes the jumpers are like the service manual configuration. You can not imagine that this error is because 2 broken resistors in the Axial board.
Thank you
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Fadhl Alakwa
GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 01, 2012 10:06
Hi
My system GE LS.X3 has the following error
"the hardware reset failed". When I replaced the STC ARTESYN III board the error disappear and I got another error when I do KV&MA (X-RAY)TEST:
"Hard time: 86098 error type:firmware software error :error when posting to vrtx queue soft or hard reset may be required."
I appreciate any advise and suggestions.
Note: I do download flash without any errors.
I am waiting your reply.
to rate and post a comment
Mark Parris
re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 02, 2012 07:24
Did you change the jumpers on the Artesyn bd?
to rate and post a comment
Fadhl Alakwa
re: re: GE LIGHTSPEED PLUS CT INITIALIZATION PROBLEM
October 04, 2012 08:44
Yes the jumpers are like the service manual configuration. You can not imagine that this error is because 2 broken resistors in the Axial board.
Thank you
to rate and post a comment