Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

GE Lightspeed 16 ICE problem

Tags: CT, GE

Anton Tonev

GE Lightspeed 16 ICE problem

August 02, 2016 10:46

We are having a problem with a GE LS16 scanner:
Durring work system started to skip the image reconstruction / no images were displayed after scan / and to "hang up".
We decided to do a fresh software installation.
We did the fresh software installation on the system which is equiped with Octane/Pegasus computer.
After this system software installation system boots and at one point when the OC is initilizing the system goes into shutdown mode and shuts down by itself.
During installation there is message "Detection of Motorola Board failed. ICE may be
down. Manually reset the VME chassis or check if /etc/hosts has all the entries.
Press OK after you are done with the reset. PLEASE NOTE: Click OK after the
reset.."
After that we are trying to test the ICE but we cannot conect to it at all with "rlogin ice" ot "ice diags or any other way to connect with the ICE.
Also we cannot ping to the ICE from console mode.
We suspect the MVME 2433 board which we had installed in the ICE but I want to ask for any ideas or experiance which you may have.
Thanks!

2 people liked this
Log inor Register

to rate and post a comment

(74)
(15)

PIEBENG APPOLIN

re: GE Lightspeed 16 ICE problem

August 03, 2016 06:01

hi
rigth now you have two problem
you have a bad ig board for before
and a timekeeper inside your MVME 2433 had gone end of it life
to confirm it try to reset the ice MVME 2433 when to the ppc prompt rb ;v

PPC1-Bug>rb ;v

good luck

1 person liked this
Log inor Register

to rate and post a comment

Anton Tonev

re: re: GE Lightspeed 16 ICE problem

August 03, 2016 09:48

Piebeng, thank you.
Yes, timekeeper is gone. I checked. 0,26 V.
Is there any way to replace this battery only?
Can I configure it?
Can you describe more detailed the test procedure?
Thank you!

Log inor Register

to rate and post a comment

(74)
(15)

PIEBENG APPOLIN

re: re: re: GE Lightspeed 16 ICE problem

August 04, 2016 11:01

can you to the ppc1-bug prompt
normaly the correct board may look like this:

PPC1-Bug>
PPC1-Bug>cnfg
Board (PWA) Serial Number = "3348651 "
Board Identifier = "MVME2306S "
Artwork (PWA) Identifier = "01-W3345F34B "
MPU Clock Speed = "300"
BUS Clock Speed = "067"
Ethernet Address = 08003E2E94F2
Primary SCSI Identifier = "07"
System Serial Number = "0000000000000000"
System Identifier = "0000000000000000000000000000000"
License Identifier = "3348651 0"

try to replace the baterie timekeeper

in application shutdown open unix shel and just reset the motorola board to get to the ppc1-bug
and then type ''cnfg'' and send me a screen picture of config
appolinpiebeng@yahoo.fr

1 person liked this
Log inor Register

to rate and post a comment

Anton Tonev

re: re: re: re: GE Lightspeed 16 ICE problem

August 04, 2016 11:31

Is this all in Unix shell?
By "just reset the Motorola board to get to ppc1-bug" you mean I have to type in something in Unix shel?
Can you tell me what?
Thanks

Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: GE Lightspeed 16 ICE problem

August 03, 2016 09:47

First stop the auto shutdown by:

Before the system startups CT applications software, a popup confirmation box is displayed.

Using the mouse, left click the [CANCEL] button within 5 seconds of the window being displayed

Immediately after clicking Cancel, a cancellation popup message box appears.

Using the mouse, left click the [OK] button.

From the TOOLCHEST, use the mouse and left click on the [UNIX SHELL] button.

At the prompt, in the Unix Shell, type: setenv NOHOSTSHUTDOWN

The above command prevents the applications startup process from shutting down if an error is encountered.

Now start applications software by typing: startup &

Applications software will startup and not terminate if an error is encountered.

1 person liked this
Log inor Register

to rate and post a comment

Anton Tonev

re: re: GE Lightspeed 16 ICE problem

August 19, 2016 08:12

Thanks

Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: GE Lightspeed 16 ICE problem

August 03, 2016 09:47

then:
Reset the VME chassis or check if .etc/hosts
has all the entries in the file.

2 people liked this
Log inor Register

to rate and post a comment

Anton Tonev

re: re: GE Lightspeed 16 ICE problem

August 04, 2016 11:28

I still cannot connect with ICE.
I think due to battery loss it forgot the configuration.
Can I reconfigure it?

Log inor Register

to rate and post a comment

Anton Tonev

re: re: GE Lightspeed 16 ICE problem

August 04, 2016 11:31

The .etc/hosts looks OK. I can see IP of the ICE configured. The thing is how to reset it after changing the battery? Thanks

Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: re: re: GE Lightspeed 16 ICE problem

August 05, 2016 10:05

{ctuser@bayxx}[1]: cd /usr/g/service/log

{ctuser@bayxx}[2]: more iceConsole.log

Comment: Begin Apps startup

Logfile updated Thu Mar 22 10:50:04 2001

"Proc_Ice" records the beginning of applications process startup in the "iceConsole log, as above.

Comment: Reset VME Chassis

[VxWorks Boot]: b

Upon executing the “b” command from boot PROM, a VME reset takes place.

Comment: VME reset successful

Copyright Motorola Inc. 1988 - 1999, All Rights Reserved

1 person liked this
Log inor Register

to rate and post a comment

CRAIG WAMBURA

re: re: re: GE Lightspeed 16 ICE problem

August 05, 2016 10:05

When the VME chassis boots up, the Motorola powerPC (RIP) logs bootup and application startup communications. The output is logged to /usr/g/service/log/iceConsole.log file. There are two methods for viewing this log. One method is using the service desktop, another is using a Unix shell and typing commands.

1 person liked this
Log inor Register

to rate and post a comment

Anton Tonev

re: re: re: re: GE Lightspeed 16 ICE problem

August 19, 2016 08:12

Thanks

Log inor Register

to rate and post a comment