60-0316H
Spit Max error.
kV control has detected the
number of tube spits during
exposure has reached the
limit (see theory of
operation, software
section)
reasonably x-ray tube -Try again at various kV/mA to
confine problem.
-Troubleshoot HV section. (x-ray
tube, cable, HV tank)
-Diagnose based on tube history.
40-0107H
Rotation Inverter
permanent overcurrent
An overcurrent has been
detected and 3 restart
have been tried
unsuccessfully within a
single rotation state
1. Rotation board
2. Tube stator winding in
short circuit -> cable
(NP) or HEMIT (only
for NP++)
3. Wiring incorrect
(shorted)
-Check wiring from rotation board
to tube.
-Troubleshoot tube windings
-Replace Rotation board
ERROR 60-0316H:
Description:kV control has detected the number of tube spits during exposure has reached the limit.
Cause: Faulty X-ray Tube
ERROR 40-0107H:
Description: An overcurrent has been detected and 3 restart have been tried unsuccessfully within a single rotation state.
Cause: Faulty Rotation Board, Tube Stator Winding short circuit.
RECOMMENDATION:
- Check cabling from rotation board to tube
- perform a NO LOAD test to isolate the cause
- troubleshoot tube windings
- Check anode HV cable
It looks like you have a shorted tube windings....and possibly a bad rotation board.
I would recommend replacement of X-ray tube and Rotation Board
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Asad
GE Hispeed CT Error Code: 40 - 0107H Rotation Error
September 21, 2020 10:37
Hello, GE hispeed NXi pro CT Scan machine When We try to calibration or Tube warmup or any scan we got the following errors:
XG Error: Phase is 5 code is 60
Detail Error Code 60-0316H
Exposure Error: Spit Max Error
we got the following error:
XG Error, Error: Phase is 2 code is 40
Detail Error Code 40-0107H
Exposure Error: Rotation error: Rotation Inverter permanent overcurrent
I will appreciate any help that could be provided.
Thank you all.
Asadullah Asad
to rate and post a comment
Dmitry Dederer
re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error
September 22, 2020 10:32
60-0316H
Spit Max error.
kV control has detected the
number of tube spits during
exposure has reached the
limit (see theory of
operation, software
section)
reasonably x-ray tube -Try again at various kV/mA to
confine problem.
-Troubleshoot HV section. (x-ray
tube, cable, HV tank)
-Diagnose based on tube history.
40-0107H
Rotation Inverter
permanent overcurrent
An overcurrent has been
detected and 3 restart
have been tried
unsuccessfully within a
single rotation state
1. Rotation board
2. Tube stator winding in
short circuit -> cable
(NP) or HEMIT (only
for NP++)
3. Wiring incorrect
(shorted)
-Check wiring from rotation board
to tube.
-Troubleshoot tube windings
-Replace Rotation board
to rate and post a comment
CRAIG WAMBURA
re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error
September 22, 2020 10:32
ERROR 60-0316H:
Description:kV control has detected the number of tube spits during exposure has reached the limit.
Cause: Faulty X-ray Tube
ERROR 40-0107H:
Description: An overcurrent has been detected and 3 restart have been tried unsuccessfully within a single rotation state.
Cause: Faulty Rotation Board, Tube Stator Winding short circuit.
RECOMMENDATION:
- Check cabling from rotation board to tube
- perform a NO LOAD test to isolate the cause
- troubleshoot tube windings
- Check anode HV cable
It looks like you have a shorted tube windings....and possibly a bad rotation board.
I would recommend replacement of X-ray tube and Rotation Board
craigwambura@outlook.com
to rate and post a comment
Maurice Dilick
re: re: GE Hispeed CT Error Code: 40 - 0107H Rotation Error
September 25, 2020 11:10
To check the Stator, measure the resistance. Normal problem of probably the cable.