Our positioning lights on the scanner could not come on and the CT refused to scan. Error message is :TGP TASK TIME OUT. Then it displays XG ERROR.
Any help is most appreciated.
your generator is detecting a problem with x-ray need to send the error that has after the xg error to determine if you have a low ma over ma, over kv or under kv problem, ir the rotor is turning, can you hear the tube prepare to do x-ray, can you hear the generator do some thing,???? please advise mor information or call us to 3033042719
Thanks Veronica.The error: 1)GANTRY ERROR CODE: 13-1002-17. Error detected by TGP(GP)driving power of servo amp is out of specification.
2)GANTRY ERROR CODE:13-1002-04 .Error detected by : TGP servo amp alarm is detected.
CC Medical provides reliable ultrasound systems, probes, system parts, & technical support serving hospitals and healthcare providers domestically & internationally. Wholesale pricing available. 833-309-4229. Se Habla Espanol sales@ccmedicalhs.com
Thomas Onyema
GE FXi ERROR CODE
July 25, 2011 08:00
Our positioning lights on the scanner could not come on and the CT refused to scan. Error message is :TGP TASK TIME OUT. Then it displays XG ERROR.
Any help is most appreciated.
to rate and post a comment
Veronica Alvarez
Re: GE FXi ERROR CODE
July 25, 2011 03:51
your generator is detecting a problem with x-ray need to send the error that has after the xg error to determine if you have a low ma over ma, over kv or under kv problem, ir the rotor is turning, can you hear the tube prepare to do x-ray, can you hear the generator do some thing,???? please advise mor information or call us to 3033042719
to rate and post a comment
Thomas Onyema
Re: GE FXi ERROR CODE
July 25, 2011 05:16
Thanks Veronica.The error: 1)GANTRY ERROR CODE: 13-1002-17. Error detected by TGP(GP)driving power of servo amp is out of specification.
2)GANTRY ERROR CODE:13-1002-04 .Error detected by : TGP servo amp alarm is detected.
to rate and post a comment
Nirvikar Yadav
Re: GE FXi ERROR CODE
July 25, 2011 07:22
Most probably Servo Amp is faulty. Contact me on +61449552069.
to rate and post a comment
Carlos Duran
Re: GE FXi ERROR CODE
August 07, 2011 08:32
You need to be clear, if the error is from XG or SVAmp.
This a very different stages of failure
to rate and post a comment
salman khan
re: Re: GE FXi ERROR CODE
January 28, 2013 04:49
Exactly Servo Amp is at fault