i am seeking for your advice about this issue...
i am getting an error of LOW INPUT POWER.when i check the error log it says SRU is LOw input power.i am wandering why it shows an error like this when we dont have a SRU. one more thing when we do the laser it doest function.i already change the axial board on the STC side and the axial encoder which are probable cause but the problem still exist...thanks for all your suggestions
I am the engineer of Santo Domingo Garcia, as is the voltage
input, the computer I spend the same and had a face of unstable voltage, you have the service manual?
CHECK phase-phase voltage and all the three phase lines i think you have only two phase available or one phase check your three phase lines and if all the line voltages are OK then you should go for taping in PDU transformer adjust the voltage acording to the main line voltage
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Marivic Mendoza
ge ct/i ct scan machine
July 04, 2012 11:59
i am seeking for your advice about this issue...
i am getting an error of LOW INPUT POWER.when i check the error log it says SRU is LOw input power.i am wandering why it shows an error like this when we dont have a SRU. one more thing when we do the laser it doest function.i already change the axial board on the STC side and the axial encoder which are probable cause but the problem still exist...thanks for all your suggestions
to rate and post a comment
karen alberto garcia salazar
re: ge ct/i ct scan machine
July 05, 2012 08:36
I am the engineer of Santo Domingo Garcia, as is the voltage
input, the computer I spend the same and had a face of unstable voltage, you have the service manual?
to rate and post a comment
TOM CHARLES
re: re: ge ct/i ct scan machine
July 06, 2012 08:40
CHECK phase-phase voltage and all the three phase lines i think you have only two phase available or one phase check your three phase lines and if all the line voltages are OK then you should go for taping in PDU transformer adjust the voltage acording to the main line voltage
to rate and post a comment
Marivic Mendoza
re: re: re: ge ct/i ct scan machine
July 09, 2012 09:00
i already check the voltages and all are present.i also adjust the tapping voltage from 480v to 380v but it still show the error low input power
to rate and post a comment
TOM CHARLES
re: re: re: re: ge ct/i ct scan machine
July 10, 2012 02:02
email me at tomcharles41@yahoo.com
to rate and post a comment
Arthur Singh
re: ge ct/i ct scan machine
October 25, 2012 08:50
CHECK THE INPUT VOLTAGES AT SERVO AMPLIFIER. MAYBE INPUT VOLTAGES ARE NOT PROPER.
to rate and post a comment
Ivan Bravo
re: ge ct/i ct scan machine
October 25, 2012 01:16
hi Marivic can you solve the problem of LOW INPUT POWER?,i have the same problem can you help me ????
to rate and post a comment
salman khan
re: re: ge ct/i ct scan machine
October 26, 2012 12:24
HELLO IVAN BRAVO
Email me at salmankhan.ge.ct@gmail.com i will help you