Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

GE CTe Dual error

CRAIG WAMBURA

GE CTe Dual error

March 02, 2015 09:13

I need assistance on the following errors that come up during calibration:
Warmup SD/Mean is out of range
warm-up HPF(ratio) is out of range.
Error in vector check.
Error in SD/Mean
Error in HPF.

Log inor Register

to rate and post a comment

(3)

(3)

Bruce Hull

re: GE CTe Dual error

March 03, 2015 10:10

This could be something as simple as having contrast on the scan window and as complex as an issue with your DAS/Detector. It makes sense to start simple. You should remove the mylar window and try your calibration again. If it still fails wipe down the detector face and inspect the collimator to make sure there is no debris or obstruction. Try your calibration again (without mylar window installed). If it still fails, we will probably need to have a phone discussion.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: GE CTe Dual error

March 03, 2015 10:10

Looks like noisy channel(s).
chech it, and clean the xray path

If need more assistance, write me

Ingenieriaclinica@yahoo.es

Log inor Register

to rate and post a comment

(32)
(4)

Tiorico_86

re: GE CTe Dual error

March 03, 2015 10:23

YOU MUST TO CLEAN THE MYLAR OR REMOVE ANY OBSTRUCTION ON THE XRAY PATH. IF YOU HAVE A CLEAN MYLAR THEN YOU MUST TO CHECK THE DETECTOR OR DAS, CHECK THE POWER SUPLIES, FUSES AND RUN DAS DIAGNOSTICS JUST TO SEE IF YOU DON'T HAVE ANY CAM BD BAD.

sysmedi22@gmail.com

Log inor Register

to rate and post a comment

(17)
(3)

RxMedicalSystemsServices

re: GE CTe Dual error

March 03, 2015 10:23

Check your bow tie filter,
Check any foreign bodies between tube and detector including collimator.
Take backup of cal files and perform calibration including Qcal.

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: GE CTe Dual error

March 03, 2015 10:28

Sorry Craig
I also got sometomes Error in HPF dirong scan or daily cal
any help?
Please explane the meaning of the error

Log inor Register