Hello Gary, I would first check your batteries to be sure one of them is not bad. Next, I would try a recalibration of the charging circuit. Finally, if those do not solve your issue replace the charging board. Hope this helps, Bill
Hello Gary,
Have you checked the plug receptacle and line cord reel?
BATTERY VOLTAGE MUST BE 113 VOLTS OR LESS TO DO LOAD TEST!
DO NOT CHARGE BATTERIES BEFORE LOAD TEST!!
a. Connect Oscilloscope From Circuit Breaker To Ground - Use Sample Mode
b. Make Exposures At 80, 100 KVP And 125 KVP, About 10 MAS
c. Read Battery Voltage And Average The Ripple At The Bottom
d. Battery Voltage Should Drop About 13 VDC +/-10% At 80 KVP
e. Battery Voltage Should Drop About 16 VDC +/-10% At 100KVP
f. Battery Voltage Should Drop About 17 VDC +/-10% At 125KVP
4. If Batteries Pass Load Test, Unit May Require Recalibration Of MAS, KVP,
Taps And Filament Current Tables
5. If Batteries Do Not Pass The Load Test, You Get To Replace Them!
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Gary Benitez
GE AMX 4 Error Code 102 - Charging
June 08, 2022 10:01
Hi Guys,
We are getting an error code of 102 related to a charging error on our GE AMX 4.
Does anyone know what part needs to be changed or re-calibrated?
Please advise.
to rate and post a comment
Raymundo Moreno
re: GE AMX 4 Error Code 102 - Charging
June 09, 2022 11:40
On charger board, check J11. On male or female connector you may observe bad contact due heathing. Fix it and recalibrate charger board.
to rate and post a comment
Okuwin
re: GE AMX 4 Error Code 102 - Charging
June 09, 2022 11:40
Hello Gary, I would first check your batteries to be sure one of them is not bad. Next, I would try a recalibration of the charging circuit. Finally, if those do not solve your issue replace the charging board. Hope this helps, Bill
to rate and post a comment
CRAIG WAMBURA
re: GE AMX 4 Error Code 102 - Charging
June 10, 2022 09:47
Hello Gary,
Have you checked the plug receptacle and line cord reel?
BATTERY VOLTAGE MUST BE 113 VOLTS OR LESS TO DO LOAD TEST!
DO NOT CHARGE BATTERIES BEFORE LOAD TEST!!
a. Connect Oscilloscope From Circuit Breaker To Ground - Use Sample Mode
b. Make Exposures At 80, 100 KVP And 125 KVP, About 10 MAS
c. Read Battery Voltage And Average The Ripple At The Bottom
d. Battery Voltage Should Drop About 13 VDC +/-10% At 80 KVP
e. Battery Voltage Should Drop About 16 VDC +/-10% At 100KVP
f. Battery Voltage Should Drop About 17 VDC +/-10% At 125KVP
4. If Batteries Pass Load Test, Unit May Require Recalibration Of MAS, KVP,
Taps And Filament Current Tables
5. If Batteries Do Not Pass The Load Test, You Get To Replace Them!
Hope this helps!
to rate and post a comment
Marco Antonio Vasquez Cortes
re: GE AMX 4 Error Code 102 - Charging
June 15, 2022 09:17
Hi Gary re-calibrated
to rate and post a comment