2.15.1 Re-exposure processing (A61-A63)If error E35 (Insufficient tube voltage), E36 (Tube over voltage), or E37 (Discharge detected) is detected, re-exposure processing will be performed up to two times. In re-exposure processing, the values of kV and mAs are the same as normal exposure processing (see the following figure).
E35
Insufficient tube voltage{MD:2.15.4}
No
- Make sure that CNKV is attached to the CPU board.- Make sure that there is no disconnection or ground fault in the tube voltage detection circuit.- Check the battery voltage during radiography.
E36
Tube over voltage{MD:2.15.3}
No
CPU board3E1-7,3E2-7: L
- Make sure that there is high-voltage discharge at the tube head. -> Check the waveforms between KV+ and KV+G (GND) and between KV- and KV-G (GND) on the CPU board.- Check whether the adjustment value of E36 is not changed to less value.
E37
Discharge detected{MD:2.15.7}
No
CPU board2D3-7: L
- Make sure that there is high-voltage discharge at the tube head.- Make sure that the main transformer is not layer shorted.- Make sure that +kV and -kV are well balanced.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
xtekvet
FCRGO A61 error
November 06, 2018 10:17
FcrGo has an A61 error message. Can still take images. Manual does not give reason for error or how to fix it. Any suggestions?
to rate and post a comment
Alfred M. Brown
re: FCRGO A61 error
November 07, 2018 08:19
2.15.1 Re-exposure processing (A61-A63)If error E35 (Insufficient tube voltage), E36 (Tube over voltage), or E37 (Discharge detected) is detected, re-exposure processing will be performed up to two times. In re-exposure processing, the values of kV and mAs are the same as normal exposure processing (see the following figure).
to rate and post a comment
Alfred M. Brown
re: FCRGO A61 error
November 07, 2018 08:19
E35
Insufficient tube voltage{MD:2.15.4}
No
- Make sure that CNKV is attached to the CPU board.- Make sure that there is no disconnection or ground fault in the tube voltage detection circuit.- Check the battery voltage during radiography.
E36
Tube over voltage{MD:2.15.3}
No
CPU board3E1-7,3E2-7: L
- Make sure that there is high-voltage discharge at the tube head. -> Check the waveforms between KV+ and KV+G (GND) and between KV- and KV-G (GND) on the CPU board.- Check whether the adjustment value of E36 is not changed to less value.
E37
Discharge detected{MD:2.15.7}
No
CPU board2D3-7: L
- Make sure that there is high-voltage discharge at the tube head.- Make sure that the main transformer is not layer shorted.- Make sure that +kV and -kV are well balanced.
to rate and post a comment