EUB-450

April 14, 2006 01:55

Dear sir,
An EUB-450 ultrasonic by Hitachi display white block.Who can give us some advice kindly or support some document?
Thank first!
htdocument@126.com

Log inor Register

to rate and post a comment

(16)

(10)

Vijai shankar raja

Re: EUB-450

April 18, 2006 07:20

Hi,
First confirm your monitor. You can give the video output from machine to another monitor and test.
regards,
Shiju
Cura Medical Equipment Pvt Ltd.
shiju@cura.co.in

Log inor Register

to rate and post a comment

Ray Coombs

Re: EUB-450

April 19, 2006 02:09

once you have verified your monitor is not the problem. if the problem is in the normal imaging scan area it is one of the DSC boards, I believe one of 2 for the EUB 450, if the problem is outside the imaging scan area look at the Video I/O Board or CPU. Hope it helps.

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

Re: EUB-450

April 24, 2006 05:46

hello:
you has a electronic board problem, located on memory image chips, and/or buffers.

you must use oscilloscope to check the data/address lines on the respective boards, to locate the fault.

regards.

Log inor Register

to rate and post a comment