Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

error voluson 730 expert

(12)
(3)

GERARDO ALCANTARA

error voluson 730 expert

January 29, 2014 08:24

Hello
I have voluson 730 expert , when I connect 4d convex probe model RAB4-8P the ultrasound give me this message: MotoCtrl error

can you help me please

Log inor Register

to rate and post a comment

Gilrubio

re: error voluson 730 expert

January 29, 2014 03:27

try with another one, probably your probe is damaged

Log inor Register

to rate and post a comment

Marcelo Antonio Correa

re: error voluson 730 expert

January 29, 2014 06:14

Hi, Gerardo!

It could the probe, as Gilrubio said, CPP or CPM (in older models) boards.

If your equipment is newer, huge chance CPP board is defective.

Hope this helps, take care, God Bless.

MAC.

2 people liked this
Log inor Register

to rate and post a comment

Ricardo Armenta

re: error voluson 730 expert

January 29, 2014 06:14

you have a problem with the transducer, is either Jamed, stucked or sensor has a problem, we can fix it, please contact me ricardo@ultrasolutions.com

Log inor Register

to rate and post a comment

(179)
(24)
(1)

Murat Canpolat

re: error voluson 730 expert

January 30, 2014 08:38

you have a problem with the transducer, I can fix it.
muratcanpolat@hotmail.com.tr

We can repair all 4D probes
Regards
Murat

Log inor Register

to rate and post a comment

(10)

ngambi joseph

re: re: error voluson 730 expert

February 05, 2014 04:02

Hi my fiend! can you tell me how to repair prob?
my mail josysony@yahoo.fr

Log inor Register

to rate and post a comment

(195)
(47)
(4)

EROTOKRITOS SKENTOS

re: error voluson 730 expert

January 30, 2014 08:38

Hallo Gerardo the problem is probably the probe but if you can test the probe in another ultrasound or other probe 4d in yours we will be certain about the problem.The other case is either the CCM board in older machines or from BT08 and newer the CRS board.We fix RAB4-8P probes is you want contact me at eskentos@gmail.com.

Log inor Register

to rate and post a comment

Robert Harris

re: error voluson 730 expert

January 30, 2014 11:44

Agreed its most likely the probe but could also be the board. We can supply new replacement boards and exchange parts

Log inor Register

to rate and post a comment

Ricardo A. Sanchez

re: error voluson 730 expert

February 19, 2014 07:12

Gerardo, did you repair the probe?

Log inor Register