Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

Erbe error code 45 and 47

Tags: Other, ERBE

(127)
(16)
(2)

bashir ahmad

Erbe error code 45 and 47

February 09, 2018 03:25

Having error 45 and 47 in Erbe 350 electrosurgery unit. any body having service manual or any help appreciated. My email is bashir30474@yahoo.com
Regards

Log inor Register

to rate and post a comment

(7)

Taheer

re: Erbe error code 45 and 47

February 10, 2018 12:05

Error 45
HF generator error during self check
HF output voltage low : Equipment defect

Error 46
Self Check
Error within the permissible limits during comparison of the power supply unit voltage with the High output voltage. Equipment defect.

Error 47
HF generator error during self check
Like 46, but error outside permissible limits. Equipment defect

Hope that helps

Regards
Taheer Butt
+92 333 4300800

Log inor Register

to rate and post a comment

(127)
(16)
(2)

bashir ahmad

re: re: Erbe error code 45 and 47

February 12, 2018 09:13

Thanks

Log inor Register

to rate and post a comment

(5)
(1)

Farooq Shah

re: re: Erbe error code 45 and 47

August 02, 2019 09:25

Hello there is Error 32 in ERBE ICC 350 Electrosurgical Unit and there is high frequency visial nise. May be in Power supply issue. Please give me solution.
Thanks
Farooq Shah

Log inor Register

to rate and post a comment

(22)
(7)

Sergey Radionov

re: re: re: Erbe error code 45 and 47

February 20, 2020 09:39

What tool do you use in your work? Are you using an adapter?

Log inor Register

to rate and post a comment

(45)
(10)

Igor69

re: Erbe error code 45 and 47

March 09, 2020 09:22

replace IC1 ICL7667 on power-board (5. board from left)

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)