Electrical interference

Tags: Ultrasound

Posted by an Anonymous User

Electrical interference

October 17, 2023 10:42

Certainly, here's the revised version:

I'm experiencing significant electrical interference in multiple hospitals when connected to AC power. However, the machine is functioning smoothly when operating on battery power. Further investigation suggests that the electrical wiring issue lies in the connection between the neutral and earth. I'm currently exploring methods to filter out the noise, having experimented with inductor coils and a line filter. My next step is to consider an isolation transformer. Any advice on this matter would be greatly appreciated.

Log inor Register

to rate and post a comment

(12)
(2)

Habib Sheikh

re: Electrical interference

October 18, 2023 10:18

Hello,

I think you have to clear the situation.
First check the interference come from ac power inlet or from circumference (environment).
As you explained, the machine works with battery without problem. So, note to AC power line.

You need a powerful ac line source, with enough ac current.
Next, a long-distance wiring and/or cable, from power source to machine is very bad. Dropped voltage and alternations comes inside. So, short wiring distance with good copper thickness is critical.

Also, with more distance from electrical power source, the noise source is more effective.
Also if you change a machine location, from a room to other room, it works good sometimes.

But first, a good earth well and a good earth connection is necessary.

Regards

Log inor Register

to rate and post a comment

(84)
(14)
(1)

sford

re: Electrical interference

October 20, 2023 05:06

confirm that the ground connection is not noisy. You can do that temporarily by using a 2-prong power plug. Why not try using a UPS?

Log inor Register

to rate and post a comment

(3)

MoniccaM

re: re: Electrical interference

October 23, 2023 09:25

Not all sites are equipped with UPS power, and for end users, a portable UPS might not be practical, especially when the equipment is mobile, making it cumbersome to move around. Additionally, due to the wiring setup at these sites, there is interference on the two-prong outlets as the earth and neutral are connected.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats
advertisement

Próximos Eventos de Subasta

C2 Management
TIMED Ending Soon
Jun 30, 2026 (173 lots)
Cbrx2sales
TIMED Ending Soon
Jul 01, 2026 (64 lots)
Medical Equipment Superstore
TIMED
Jul 14, 2026 (377 lots)