Over 450 Pennsylvania Auctions End Today - Bid Now
Over 1650 Total Lots Up For Auction at Five Locations - Over 100 Missouri Auctions End Tomorrow 07/15 - Bid Now, NJ Cleansweep 07/16, CA 07/17, VA 07/21, CA 07/22

Drystar AXYS printer issue

Tags: Mammo, HOLOGIC

(1)

Heather Carlson

Drystar AXYS printer issue

May 11, 2015 04:02

My facility has a Hologic Selenia Dimensions Mammo unit and a Drystar AXYS printer with 8x10 and 10x12 film. We had temporarily ran out of 8x10 film so were printer on 10x12 only. We did this by unchecking the auto print size box on the Hologic unit and manually checking the 10x12 box. Upon getting new 8x10 film the printer will only print on 10x12 and the QC values have changed. Even if I select 8x10 it prints on a 10x12. I've spoken with Hologic and they have assured me that it must be a printer issue. I've tried recalibrating the film, changing the tray sizes, turning the unit off, etc. Nothing has worked so far.

Log inor Register

to rate and post a comment

(2681)
(674)
(21)

Lenny Place

re: Drystar AXYS printer issue

May 12, 2015 08:32

I have seen this same issue and it turned out to be the Hologic if you have a system backup restore it. For some odd reason when you unchecked the auto print function it disabled the film size auto detect and is using the 10x12 as default

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED Ending Soon
Jul 14, 2026 (498 lots)
Cbrx2sales
TIMED Ending Soon
Jul 15, 2026 (124 lots)
PBC Medical Solutions
TIMED
Jul 17, 2026 (11 lots)
C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (135 lots)