CT/e single slice

December 16, 2013 04:25

Hi every one

Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice



Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: CT/e single slice

December 16, 2013 07:07

If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: CT/e single slice

December 17, 2013 08:27

Thank you for reply.
The connection to panel has 4 lines , please tell me which lines to be short-connected .

Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: re: re: CT/e single slice

December 17, 2013 01:26

There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com

Log inor Register

to rate and post a comment

(57)
(3)
(6)

R.K.S singh

re: CT/e single slice

December 17, 2013 08:28

please email me at rohitsingh4u@hotmail.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Medical Equipment Superstore Surplus Auctions July 14th

Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)