Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice
If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.
There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com
Medical Equipment Superstore closeout auctions take place on DOTmed.com. Bid today on 276 lots of medical equipment. Items can be inspected in person or via video call. Contact Rick at +1 (215) 669-3913 or info@medicalequipmentsuperstore.com
CT/e single slice
December 16, 2013 04:25
Hi every one
Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice
to rate and post a comment
Lin Dong
re: CT/e single slice
December 16, 2013 07:07
If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.
to rate and post a comment
Galal
re: re: CT/e single slice
December 17, 2013 08:27
Thank you for reply.
The connection to panel has 4 lines , please tell me which lines to be short-connected .
to rate and post a comment
Lin Dong
re: re: re: CT/e single slice
December 17, 2013 01:26
There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com
to rate and post a comment
R.K.S singh
re: CT/e single slice
December 17, 2013 08:28
please email me at rohitsingh4u@hotmail.com
to rate and post a comment