Your Email:
kevin mattingly
bv pulsera c-arm
October 06, 2020 10:16
bv pulsera giving errors m208 and m343. Info says problem with sha4 pcb or cabling. is cabling the interconnect cable or another cable.
to rate and post a comment
Herrmann
re: bv pulsera c-arm
October 07, 2020 11:13
Hi, Have you this Errors by high KV? Maybe by lateral Position of the C-Arm? KR Willi
rockbob
October 12, 2020 09:50
check for fan spinning and for dust or debris blocking opto that monitors fan spinning
tim cowell
October 14, 2020 10:03
If you cannot get a successful performance test And you are getting these errors. The most likely cause is gonna be the x-ray tube.
50+ Stretchers in stock for immediate rentals. Contact us for a today for a quote.
kevin mattingly
bv pulsera c-arm
October 06, 2020 10:16
bv pulsera giving errors m208 and m343. Info says problem with sha4 pcb or cabling. is cabling the interconnect cable or another cable.
to rate and post a comment
Herrmann
re: bv pulsera c-arm
October 07, 2020 11:13
Hi,
Have you this Errors by high KV? Maybe by lateral Position of the C-Arm?
KR Willi
to rate and post a comment
rockbob
re: bv pulsera c-arm
October 12, 2020 09:50
check for fan spinning and for dust or debris blocking opto that monitors fan spinning
to rate and post a comment
tim cowell
re: bv pulsera c-arm
October 14, 2020 10:03
If you cannot get a successful performance test And you are getting these errors. The most likely cause is gonna be the x-ray tube.
to rate and post a comment