Bennett contour error message.
coll and Relay
dont have books, anyone know what error codes these are or any direction to go.
thanks

Log inor Register

to rate and post a comment

(38)
(1)
(16)

Felix Donoghue

Re: Bennett Contour error message

April 08, 2006 02:53

I have a complete service manual, if you need it i can copy and send you one.
Thanks

Log inor Register

to rate and post a comment

(11)
(5)

edward bartra

Re: Bennett Contour error message

April 08, 2006 12:22

thank you but we will be getting a service manual. we need to get the machine running to shaow to a customer and with these two errors it won't work. is anyone familiar with this machine?????

Log inor Register

to rate and post a comment

(110)
(10)
(3)

Deon Armes

Re: Bennett Contour error message

April 09, 2006 12:46

We were a Bennett Dealer for five years (94-99)in Tampa. The fact that you have two errors from totaly different places on the machine tells me there is possibly a bad PIA chip on the CPU PCB they are 6829 chips on CPU PCB. The coll error usually says the auto collimator is not at the right size (stuck). The relay error says the backup contactor did not come in upon prep. I would swap the PIA chips on the CPU PCB and see if the errors change or you have different problems. They are in sockets so this is easy. If you need a schematic for the CPU PCB call me I can fax it to you.
Good luck!
Deon

Log inor Register

to rate and post a comment

stats
DOTmed text ad

GE, Siemens, Philips, & Toshiba/Canon imaging parts, training & support!

DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com

DOTmed Certified stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)