Over 800 Total Lots Up For Auction at Four Locations - VA 07/21, CA 07/22, KS 07/24, CA 07/31

ATL 5000 error # 002

Tags: Ultrasound, ATL

muhammad shahid

ATL 5000 error # 002

September 15, 2014 09:16

machine reboot and given this error someone help of this error. thanx and best reqards

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: ATL 5000 error # 002

September 15, 2014 06:19

removed the probes from the system, specially if you have a L12-5 and reboot the system, then plug the lineal probe in the center connector and try. for more help contact me rey@allultrasoundservices.com

Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: ATL 5000 error # 002

September 15, 2014 06:22

"0002 FEC alert message -
The front end has stopped scanning because of some power monitor problem. This could be a hardware problem with the scanhead, the channel boards, the AIM, or the power supplies, but this also could be a system software problem. In reviewing the previous error messages, from the FEC, the board errorID should indicate if there is a specific problem with hardware."

Manuals to your scanner are available on the Internet ;-). Look for it as PHILIPS HDI-5000. Google to you in the help.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: ZRG Medical Surplus Auctions July 31st

ZRG Medical closeout auctions take place on DOTmed.com. Bid today on 180 lots of medical equipment. Inspection is by appointment only. Please contact 760-438-8835 ext. 100 for an appointment.

DOTmed Certified DOTmed 100 stats
advertisement

Próximos Eventos de Subasta

C2 Management
TIMED
Jul 21, 2026 (345 lots)
Surplus Service
TIMED
Jul 22, 2026 (132 lots)
Red Kingdom Wholesale
TIMED
Jul 24, 2026 (115 lots)
ZRG Medical
TIMED
Jul 31, 2026 (183 lots)