Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
evg evgen
Aquilion32 Help!
April 14, 2017 05:50
Hello!
Help solve the problem with the crescent-shaped artifact on Aquilion32
On the scan it is presented in the form of a white strip.
Appears after some time of work.
If you turn off the computer and turn it on again, the artifact disappears
to rate and post a comment
QuantaMed
re: Aquilion32 Help!
April 17, 2017 04:43
Please did you check DAS power supply- Detector Temperature- Air calibration must not be old. Please can you test both reconstruction rack ? Also check if a red led lit on SCRT or GCIF PWB when artifact appears ( this system is MUDAT or SureCom?
to rate and post a comment