Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

AMX 4 plus error code 45

(7)

DELMAR CHAKAWA

AMX 4 plus error code 45

November 15, 2009 01:05

Unit intermittently exposes OK and may not expose during the next attempt. It then registers error 45 and declares low battery on the display window. During 'prep' the battery voltage drops from 113V to 110V. The battery stays around 109V - 110V during an exposure. It appears that the batteries are OK. What would be the next area to check?

Log inor Register

to rate and post a comment

(52)
(5)
(8)

Samuel Sandlin

Re: AMX 4 plus error code 45

November 15, 2009 05:50

You are dropping too much battery voltage during prep. Should drop 2 at maximum. Replace the entire battery set. sam@amxrayservice.com

Log inor Register

to rate and post a comment

(7)

DELMAR CHAKAWA

Re: AMX 4 plus error code 45

November 15, 2009 06:36

Correct me if I am wrong. I understand that if I drop by more than 17V at full exposure then I should change batteries. My batteries were only changed 1 year ago. I will compare this voltage drop with one of our good working unit.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: AMX 4 plus error code 45

November 16, 2009 10:30

It may be that you have loose connections on the battery set if they were just replaced one year ago.
If the batteries are older they should be able to hold a charge, one simple test is to turn unit on and turn the light on repeatedly (10 mins or so, light on every two mins.), the bar graph should not go down. if it goes down then check battery capacity.
May need to re-cal the battery voltage feedback, you'll have to do a complete calibration afterwards but it's only 2-3 hours work and doen't cost much.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)