Again, CT/ e system problem

Posted by an Anonymous User

Again, CT/ e system problem

April 28, 2014 08:54

The problem :

Both HDDs are damaged ,so there is no booting of software at all .
I changed both HDDs and made LFC procedure ; I loaded the OS ( CTT 4.5.1) and installed the application SW ( 3.12 ).
After rebooting ,I wanted to install patches ,so I inserted the CD in DVD RAM and clicked on INSTALL SOFTWARE , a window appears " do you want to load from DVD " I clicked on YES . After 10 seconds a window appears says " Timeout.... check CD-ROM drive "
Dismiss .

please note that DVD RAM is configured and it is installed in place of CD-ROM and MOD in the system .

So , how /why the system dose not detect the DVD-RAM drive ??

Another problem :
SYSTEM dose not start up , because HINV ERROR appears as power test failer . i think that is because DBPCI is not detected by system .

I can not install patches ,which are important at this stage , because the application software dose not find the DVD-RAM drive !!!

Please help

Log inor Register

to rate and post a comment

(181)
(27)

Tibor Ivanszky

re: Again, CT/ e system problem

April 29, 2014 08:46

If device is configured in the OS does not mean the application will recognize it I guess.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)