Over 1650 Total Lots Up For Auction at Two Locations - PA 07/14, NJ Cleansweep 07/16

AC-3, error 03c2

Tags: PACS / DR / CR / Digital, FUJI

(28)
(4)

MOI

AC-3, error 03c2

January 23, 2014 12:10

Hello team technicians, please can you help me in FUJI AC-3, i have error 03c2 , i have service manual, but i don't find problem, maybe SZ1 but where is it ??? thanks your help in advance.
Regards
rdus0001@yahoo.fr

Log inor Register

to rate and post a comment

(9)
(4)
(5)

Indarpersad Singh

re: AC-3, error 03c2

January 24, 2014 08:19

Look at page 416 of your manual. Parts List for sub scanning Unit for location of SZ1. It is #23

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Source for Sony Medical Displays, Printers & More!

Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.

stats

Próximos Eventos de Subasta

Medical Equipment Superstore
TIMED
Jul 14, 2026 (502 lots)