Over 100 Kansas Auctions End Tomorrow 07/24 - Bid Now
Over 700 Total Lots Up For Auction at Three Locations - TX 07/30, CA 07/31, MO 08/04

16 slice Lightspeed problem

(183)
(13)
(2)

Borhan Kalash

16 slice Lightspeed problem

October 02, 2009 04:11

I have a trouble in the 16 slice Lightspeed hope sombody can help me:
when we request warm up the system did not finish prepration and the it shows the following 2 error:
Error : 185557
could not send HEMRC message in the allotted time.
HEMRC serial link is down.
HEMRC_ISO_+12V failure detected posible cause :HEMRC fuse (F5) interconnects,HEMRC AC Drive,HEMRC control board ....

Error: 183334
HEMRC fault could not be reset.


I found that the 2 fuses F1 and F2 on the HEMRC Interface Board are burned.

I replaced the 2 fuses and tried to do a lot of scans but i still got the same errors.

note that the 2 fused did not burn again.

note that during the scan tried there was not a HV on the PDU (the resonsable contactor did not switch to ON).

What do you think the cause of the a.m. trouble?

Regards.



Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: 16 slice Lightspeed problem

October 03, 2009 08:10

Do you check 12vdc after change fuses, isn't it?

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

Re: 16 slice Lightspeed problem

October 05, 2009 09:19

Dear Jose,
thanks, where should I find this 12V?
Regards

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: 16 slice Lightspeed problem

October 06, 2009 09:25

Hi Borham:

Really I don't know. I am suggesting you to check by logic. Sometimes fuses blow but damaged is done. Error log say "lost 12vdc soo try to find board or section named at error and check low voltajes power supply. You can follow the original blow fuses to determinate what caused the short circuit. PROBABLY THE CAUSE IS STILL THERE but there is not any voltage due any component is open like the 12vdc regulator

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Cyclic Materials - Sell us your permanent magnet MRIs! WE BUY WARM MAGNETS FOR CASH

Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621

stats

Próximos Eventos de Subasta

Red Kingdom Wholesale
TIMED Ending Soon
Jul 24, 2026 (115 lots)
DeoCorp
TIMED
Jul 30, 2026 (229 lots)
ZRG Medical
TIMED
Jul 31, 2026 (229 lots)
Cbrx2sales
TIMED
Aug 04, 2026 (272 lots)