BONE DENSITOMETER HOLOGIC EXPLORER INDEX 298

Tags: Bone Densitometer, HOLOGIC

(3)

Benji Iglesias

BONE DENSITOMETER HOLOGIC EXPLORER INDEX 298

April 15, 2024 10:38

Hi colleagues,

Wanted to ask about a problem I have with a densitometer Hologic explorer, I've already change X ray controller board, C-arm a
Board and look the voltages of LVPS +24, +15, -15.

The system connects and the application runs without any problem. When it's time make an exposure is when the problem occurs.

On QC the system position itself the laser turns on and then when it's time to exposure it gives the error. Even on x ray survey there is no exposure

I will appreciate any suggestions.

Thank you

Log inor Register

to rate and post a comment

(53)
(5)

kco

re: BONE DENSITOMETER HOLOGIC EXPLORER INDEX 298

April 16, 2024 02:41

Good morning Benji, I had a similar experience to yours with the quality test, after much review we found the problem in the rx output window, for some reason there was dielectric oil in this area, it seems that this caused the error, clean and put it back together.
The equipment needs to pass the quality test to work.

Log inor Register

to rate and post a comment

(21)

(1)

AKIS PAPAS

re: BONE DENSITOMETER HOLOGIC EXPLORER INDEX 298

April 17, 2024 11:28

Oil leak in the colimator will not prevent the X-Ray survey utility from turning on X-Rays! KCO's reply does not apply in your case. What is the error's description that you see? Have you confirmed that the filter drum is corectly aligned and has good timing belts before checking anything else?

Log inor Register