Siemens Harmony Problem
December 07, 2016 10:19
Hello friends ,
When the Syngo start up at a Siemens Harmony MRI , a window appears :
MrVCheck , inform the local service ,Findings occurred check the logfile :
c:\MedCom\log\MrVCheck.log
it counts till 99 and then it logs into syngo software succesfully.
what is the problem??
thanks in advance
to rate and post a comment
Marshall Shannon
re: Siemens Harmony Problem
December 08, 2016 09:21
c:\MedCom\log\MrVCheck.log
What does the MrVcheck.log say?
It should state in the LOG the issues. Did you read the log?
to rate and post a comment
Gary Provenzano
re: Siemens Harmony Problem
December 08, 2016 09:23
Don't delete the log file....delete medcom/config/MRI/MrSystem MrVCheck.ref
to rate and post a comment
Gary Provenzano
re: Siemens Harmony Problem
December 08, 2016 09:23
Don't delete the log file....delete medcom/config/MRI/MrSystem MrVCheck.ref
to rate and post a comment
Airat Saitov
re: Siemens Harmony Problem
December 08, 2016 09:24
Hello, you will find file MrVCheck.ref in C:\MedCom\config\mri\mrsystems and rename it or delete. It will fix this warning.
to rate and post a comment
kosale
re: Siemens Harmony Problem
December 09, 2016 08:50
OK thanks :-)