Siemens Somatom Sensation 64 Error Gantry Cover Open

February 05, 2020 10:20

Hello,

Anyone who can help me on rectifying the error Gantry Cover Open. We checked the 2 switch for the gantry cover and it is working well. We disconnect the switch and Switch ON the system. During Calibration and checkup the error still appears.

Your reply is most appreciated.

Thank you

1 person liked this
Log inor Register

to rate and post a comment

(28)
(5)

Shawn McNeir

re: Siemens Somatom Sensation 64 Error Gantry Cover Open

February 06, 2020 10:32

It appears one of your cover switches are not working or your cover isn’t hitting the switch correctly.

You said you removed the switch, but did you short the contacts to simulate the switch closed?

1 person liked this
Log inor Register

to rate and post a comment

(17)
(2)

Salvador III Gaviola

re: re: Siemens Somatom Sensation 64 Error Gantry Cover Open

February 07, 2020 10:23

Hello Mr. Shawn,

No. We remain disconnected but still error appears. What board is this switch are connected? Is it possible that the board is defective?

in addition to that, the switch is normally closed. When you closed the gantry cover, the switch will open.

Any further advises.

Thank you

Log inor Register

to rate and post a comment

(305)
(56)

Gabriel Cool

re: Siemens Somatom Sensation 64 Error Gantry Cover Open

February 06, 2020 10:33

You can't work without disconnected switchs. Push on switch by hand and check error

Log inor Register

to rate and post a comment

(17)
(2)

Salvador III Gaviola

re: re: Siemens Somatom Sensation 64 Error Gantry Cover Open

February 07, 2020 10:23

Hello Gabriel,

The error appears during calibration and checkup. When you press the Move button in the control panel, it will proceed to the next procedure. During the checkup and calibration, it appears once in every procedure.

Thank you

Log inor Register

to rate and post a comment

(17)
(2)

Salvador III Gaviola

re: re: re: Siemens Somatom Sensation 64 Error Gantry Cover Open

February 10, 2020 09:24

We rectify the error/warning.

Thank you