Rotor Will Not Stop Turning Unless The Batteries Are Unplugged - Técnicos de mantenimiento Foro

Rotor will not stop turning unless the batteries are unplugged

Tags: C-Arm, OEC

(13)

Keith Lartigue

Rotor will not stop turning unless the batteries are unplugged

June 20, 2019 09:44

Hello my fellow Engineers,

I have a 9800 Plus c-arm that has a rotor that constantly spins and the only way to stop it, is by unplugging the batteries or disconcerting the tube connector. Please help! I've replaced the batteries and xray tube already.

Log inor Register

to rate and post a comment

(82)
(7)
(6)

Alfred M. Brown

re: Rotor will not stop turning unless the batteries are unplugged

June 21, 2019 07:57

Check X-ray control PCB 00-879803 IC U22 and U23. also check Power/Motor Relay PCB 00-879323.

Log inor Register

to rate and post a comment

(354)
(34)
(4)

Juan de la Cruz Martinez

re: Rotor will not stop turning unless the batteries are unplugged

June 21, 2019 07:58

I don't see how the batteries or the tube could be the problem? Do you have a manual, or knowledge of this Carm?. Look for what drives the rotor (rotor control , or similar circuit). Let me know, I can look on the schematics if you need further help.